We have twenty 2116e and we want to calibrate them, but I couldn't find the "cal" menu, in the user manual chapter 5.5 there is a navigation diagram, I followed the steps but the menu skips "Cal" and jumps to "AL"
could you help me with this issue?
Thanks!!
Help get this topic noticed by sharing it on
Twitter,
Facebook, or email.


-
Hello Efrain,
There is no CAL menu in the 2116e, do you mean 2216E?
If it is the 2216e then this can be found in the Config menu. -
-
we have both model, but if there is no cal menu in 2116e, how can we calibrate this equipment? I mean, we have a process calibrator to verify or compare both readings, the 2116e display and my process calibrator, if the diference is <= 0.25% the verification is good, but what happen if the value is greater?
is there an option to match this values?
thanks for your support! -
-
-
-
Oops!...Nope, I said I got it because I found the right user manual but when I checked in the instrument 2116e, the input menu only showed me Fil, Cjc y Mv.
-